We developed a micro-magnetometry with a 2.5 μm spatial resolution based on micro X-ray magnetic circular dichroism (XMCD) technique in order to study magnetic properties of dot arrays for bit-patterned media. This micro-magnetometer was applied to the magnetic characterization of Co–Pt dot arrays fabricated by ion beam etching. As the dot size became small, the intensity of XMCD drastically decreased for dots fabricated by Ga-focused ion beam. This suggested that the dot edges were damaged magnetically by implantation of Ga ions. The damaged width of the dot edge was estimated to be about 13 nm from the decrease in XMCD intensities. This damaged edge width agreed with the ion-implanted area estimated by Monte-Carlo simulation. The less-damaged effect of Ar ion etching was verified by the XMCD measurement of Co–Pt dots with diameter of 20 and 70 nm. It was concluded that ions with inertness, lower energy and smaller atomic number should be used to fabricate dot arrays with an areal density of 1 Tbit/in 2.