Low pressure drop thermal Mass Flow Controllers are generally thought to fulfill needs concerning the realization of a dynamic reference gas mixture generator for accurate gas analysis. A small air flow rate at low pressure drop must be controlled in a stable and precise way in the generator. True operative pressure drop limits, set point reproducibility, calibration needs and flow rate stability during operations were investigated for a low pressure drop thermal Mass Flow Controller. The flow rate bias due to late calibration and flow rate short-term stability were measured and discussed. The Allan method was used to calculate stability during operation. Calibration uncertainty, bias for late calibration, stability and set point reproducibility were composed to calculate the total uncertainty of the flow rate as a function of the operation time. Results show that it is possible to operate below the target uncertainty stated for a dynamic generator of gas mixtures down to 100Pa pressure drop. Stability gives the main contribution to total uncertainty at very short operation times, while calibration uncertainty gives the main contribution to total uncertainty at normal operation times. The calibration uncertainty at 0.1% is low enough to assure the target uncertainty for operation times over 10s. Daily verification of calibration enhances the reliability of the measurement. An accurate voltmeter is necessary for the reproducibility of the set point.
Read full abstract