A rise time study of mercuric iodide detectors irradiated with gamma ray enables one to identify the incluence of several trapping centers on the detection performance of the devices. A careful analysis of the rise time spectra identifies some of the dominant traps as having detrapping times of 300 and 900 ns, and a less intense one as having a detrapping time of 200 ns. These findings have been used to get better performance from the detectors. Discriminating those parts of the pulses that have been subjected to these trapping centers dramatically improves detection performance, especially at high energies. Tuning the electronics to avoid statistical noise due to these traps also improves performance, as does cooling the detector to about 0°C, thus “freezing” the trapping centers and producing better energy resolution. These modes of operation allow thicker detectors to be used for high energies, still maintaining good performance. At low energies (∼- 6 keV), much better energy resolution was obtained. Polarization effects were eliminated, and lower backgrounds and more symmetric peaks were achieved.