Introduction: Ionizing radiation is applied in various fields, and dosimetric control guarantees the quality and safety of products during the irradiation process. There is a need for dosimetric calibration procedures in industrial irradiation plants, specifically for self-shielded irradiation systems. Objective: To implement high dose dosimetry in the routine control of irradiated materials within a self-shielded irradiator. Methods: We worked with 32 dosimeters; we divided the dose range used into 8 points and used four RED Perspex dosimeters per point. We measured the specific absorbance for each dosimeter; plotted these values against dose and produced a third-degree polynomial fit as a calibration curve. Results: We obtained the calibration curve with an r2 of 0,9997. The uncertainties due to the dispersion of the dosimeters and the calibration curve were 1,39% and 0,22%, respectively, for a total uncertainty of 4,80%. This uncertainty includes dose determination with a factor coverage (k) equal to 2 for a 95% confidence interval. Conclusion: Perspex RED dosimeters can be used for routine control of irradiated products in a self-shielded irradiator system.