BackgroundThyroid dose reduction during brain CT scan procedures, using protective lead clothing such as lead thyroid collars were approved in the last decades. Recently, investigators have focused on applying and designing light and non-toxic shields that contain no lead. ObjectiveThis study aimed to compare the efficiency of non-lead and lead thyroid shields in radiation protection during brain CT scans based on Monte Carlo (MC) simulation and experimental measurements. MethodsThis study was performed in two steps. The first step was the MC simulation of a different combination of shielding materials to assess the efficiency of X-ray attenuation in these shields. The second step was the construction and testing of the shield, which was more efficient in MC calculation as a thyroid shield in brain CT scan. The absorbed dose of a Rando phantom thyroid was determined by both non-lead and lead shields as thyroid protectors. The thyroid radiation dose was measured with and without the usage of radiation shields. The percentage difference between the dose in the presence of the shield and without the shield was calculated. ResultThe percentage difference of the surface radiation doses on thyroid location with non-lead (combination of 45% Tungsten- 55% Tin) and lead shields in comparison to without application of the radiation shield were 35.88% and 20.73, respectively. The dose reduction at the depth of the thyroid was 39.79% and 21.49% applying non-lead and lead shields, respectively. ConclusionAccording to the results, the use of non-lead shield can reduce the absorbed dose of thyroid more efficiently than the lead shields.