Creating a high-quality image with the low patient dose is one of the most important goals in medical X-ray imaging. In this study, the image quality parameters of the digital radiographic units in Tabriz city were considered and compared with the international protocols. The image quality parameters were measured at 11 high workload digital radiography (DR) imaging centers in Tabriz city, and the results were compared to DINN 6868/58 standards. All centers equipped with the direct DR units passed the spatial resolution, low contrast detectability, contrast dynamic range, and noise tests, while the computed radiography (CR) units only could pass the two last tests. The highest spatial resolution was observed 3.2 lp/mm in the DR unit while the lowest one was 1.8 lp/mm in the CR unit. The highest noise was measured to be 0.03 OD that was observed in the DR unit. The most difference between the nominal and measured peak kilovoltage and mAs was 3.1% and 6.8%, respectively. The entrance surface air kerma in all units was obtained <0.63 mGy. The measured half-value layer range was between 2.4 and 3.54 mmAl. The physical parameters of image quality such as spatial resolution, contrast, and noise are robustness quantitative parameters for the assessment of the image quality performance of the units. Therefore, measurement and control of these parameters using two-dimensional phantoms are very critical.
Read full abstract