Transcranial focused ultrasound has been studied for non-invasive and localized treatment of many brain diseases. The biggest challenge for focusing ultrasound onto the brain is the skull, which attenuates ultrasound and changes its propagation direction, leading to pressure drop, focus shift, and defocusing. We presented an optimization algorithm which automatically found the optimal location for placing a single-element focused transducer. At this optimal location, the focus shift was in an acceptable range and the ultrasound was tightly focused. The algorithm simulated the beam profiles of placing the transducer at different locations and compared the results. Locations with a normalized peak-negative pressure (PNP) above threshold were first found. Then, the optimal location was identified as the location with the smallest focal volume. The optimal location found in this study had a normalized PNP of 0.966 and a focal volume of 6.8% smaller than without the skull. A Zeta navigation system was used to automatically place the transducer and track the error caused by movement. These results demonstrated that the algorithm could find the optimal transducer location to avoid large focus shift and defocusing. With the Zeta navigation system, our algorithm can help to make transcranial focused ultrasound treatment safer and more successful.