PurposeThis study is about the development of a new dual wavelength reading method of Fricke-Xylenol orange-Gelatin (FXG) gel dosimeters on the Vista16™ optical Computed Tomography (CT) scanner to perform 3D dose distribution measurements in stereotactic and dynamic radiotherapy treatments. MethodsThe dosimetric characteristics of an optimized FXG gel composition and its optical CT readout have been evaluated. A dual wavelength reading method has been developed on the CT scanner at wavelengths 590 nm and 633 nm. Small-field dose profile measurements with FXG gel and microDiamond (PTW) detectors were compared by γ-index analysis (0.5%/0.5 mm) to validate this method. ResultsThis reading method exhibits linear calibration curves in the 0–4 Gy and 2–10 Gy dose ranges at 590 nm and 633 nm respectively. The absorbed dose values below 4 Gy, measured at 590 nm, and those above 4 Gy, measured at 633 nm, are combined to plot a complete profile. A γ passing rate of 93.4% was achieved. ConclusionsThe new reading method of FXG gel dosimeters has been implemented on the Vista16™ scanner to span absorbed doses representative of stereotactic and dynamic radiotherapy treatments and enable 3D measurements in tumor volumes and surrounding healthy tissues. Small-field profile measurements validated this reading method as FXG gel dosimeters and microDiamond detectors were in very close agreement. This dosimetric method is a promising candidate for 3D quality assurance end-to-end tests in stereotactic and dynamic radiotherapy.