Abstract A new flow injection system for the determination of mercury by the cold vapor atomic fluorescence method is described. A sample solution (64 μ1) is injected into a stream of 0.1 M hydrochloric acid, which is mixed with a stream of 3% tin (II) chloride solution in a mixing joint. The combined stream is carried through a reaction coil for reduction of Hg (II) to Hg (0) and subsequently introduced into a specially designed gas-liquid separation vessel. Then the vaporized mercury is swept into a flow type fluorescence cell with a continuous flow of argon after removal of water in the gas phase through a condenser. Mercury is excited with an electrodeless discharge lamp as a source and the mercury fluorescence at both 184.9 and 253.7 nm is measured with a solar-blind photomultiplier. This method allows about 35 determinations of mercury in aqueous samples per hour. The calibration curve is linear over the 0–20 ppb range of mercury. The limit of detection (S/N = 3) is 0.008 ng (0.12 ppb × 64 μ1) and t...