PurposeAn ocular applicator that fits a commercial proton snout with an upstream range shifter to allow for treatments with sharp lateral penumbra is described. Materials and MethodsThe validation of the ocular applicator consisted of a comparison of range, depth doses (Bragg peaks and spread out Bragg peaks), point doses, and 2-D lateral profiles. Measurements were made for three field sizes, 1.5, 2, and 3 cm, resulting in 15 beams. Distal and lateral penumbras were simulated in the treatment planning system for seven range-modulation combinations for beams typical of ocular treatments and a field size of 1.5 cm, and penumbra values were compared to published literature. ResultsAll the range errors were within 0.5 mm. The maximum averaged local dose differences for Bragg peaks and SOBPs were 2.6% and 1.1%, respectively. All the 30 measured point doses were within +/-3% of the calculated. The measured lateral profiles, analyzed through gamma index analysis and compared to the simulated, had pass rates greater than 96% for all the planes. The lateral penumbra increased linearly with depth, from 1.4 mm at 1 cm depth to 2.5 mm at 4 cm depth. The distal penumbra ranged from 3.6 to 4.4 mm and increased linearly with the range. The treatment time for a single 10 Gy (RBE) fractional dose ranged from 30 to 120 s, depending on the shape and size of the target. ConclusionsThe ocular applicator's modified design allows lateral penumbra similar to dedicated ocular beamlines while enabling planners to use modern treatment tools such as Monte Carlo and full CT-based planning with increased flexibility in beam placement.