A capillary zone electrophoretic (CZE) method was developed, validated, and applied for the assay of metformin (MET) and pioglitazone (PIO) in pharmaceutical formulations. The optimum running buffer composition was found to be 75 mmol/L phosphate buffer containing 30% acetonitrile (ACN) at pH 4.0. The optimum instrumental conditions were found to be injection time, 10 s; applied voltage, 25 kV; hydrodynamic injection pressure, 0.5 psi for 10 s, capillary temperature, 25 °C; and the detection wavelength, 210 nm. The quantifications were calculated based on the ratio of the peak areas of analytes to atenolol as an internal standard. The CZE method was validated in terms of accuracy (98.21-104.81%), intra- and inter-day precision of migration time and peak area (relative standard deviation ≤ 5%), linearity (correlation coefficients ≥ 0.9985), limit of detection (≤0.277 μg/mL), and limit of quantitation (≤0.315 μg/mL). The proposed method was applied for the analysis of PIO and MET both individually and in a combined dosage tablet formulation. All electrophoretic parameters were calculated and evaluated. A previously reported high-performance liquid chromatographic (HPLC) method was also applied to the same samples. A comprehensive comparison was then carried out for the analytical features of both methods CZE and HPLC. Comparable results were obtained with the advantage of reagent consumption and separation efficiency of CZE over HPLC and shorter analysis time by HPLC compared with CZE.