PurposeThe efficacy of MR-guided radiotherapy on a MR-LINAC (MR-L) is dependent on the geometric accuracy of its MR images over clinically relevant Fields-of-View (FOVs). Our objectives were to: evaluate gradient non-linearity (GNL) on the Elekta Unity MR-L across time via 76 weekly measurements of 3D-distortion over concentrically larger diameter spherical volumes (DSVs); quantify distortion measurement error; and assess the temporal stability of spatial distortion using statistical process control (SPC). MethodsMR-image distortion was assessed using a large-FOV 3D-phantom containing 1932 markers embedded in seven parallel plates, spaced 25 mm × 25 mm in- and 55 mm through-plane. Automatically analyzed T1 images yielded distortions in 200, 300, 400 and 500 mm concentric DSVs. Distortion measurement error was evaluated using median absolute difference analysis of imaging repeatability tests. ResultsOver the measurement period absolute time-averaged distortion varied between: dr = 0.30 – 0.49 mm, 0.53 – 0.80 mm, 1.0 – 1.4 mm and 2.28 – 2.37 mm, for DSVs 200, 300, 400 and 500 mm at the 98th percentile level. Repeatability tests showed that imaging/repositioning introduces negligible error: mean ≤ 0.02 mm (max ≤ 0.3 mm). SPC analysis showed image distortion was stable across all DSVs; however, noticeable changes in GNL were observed following servicing at the one-year mark. ConclusionsImage distortion on the MR-L is in the sub-millimeter range for DSVs ≤ 300 mm and stable across time, with SPC analysis indicating all measurements remain within control for each DSV.