BackgroundThe study objective was to validate the relative biological effectiveness (RBE) in RayStation for carbon-ion radiotherapy (CIRT) using the Syngo treatment planning system as reference.MethodsLocal effect model I was established in RayStation (Ray-LEM) with the same parameters as in LEM I in Syngo (Syngo-LEM). Three cube plans covering most of the tumors treated at our center were generated with Syngo-LEM. Ray-LEM re-calculated the Syngo plans and compared the RBEs to the Syngo counterparts. The results showed that RayStation RBE was smaller than Syngo RBE. To ensure that Ray-LEM reproduced Syngo RBE, the observed deviations were used to scale the maximum RBE (RBEmax) in Ray-LEM. After this calibration, we further compared the RayStation RBE to Syngo RBE using additional plans in both homogeneous phantoms and patients, to ensure that the calibrated Ray-LEM reproduced Syngo RBE even with more complex planning features.ResultsThe calibration increased the RBEmax by 2.3% to raise the Ray-LEM RBE. The target mean RBE deviations in the phantom evaluation plans were median: 0.0 (minimum: − 1.1 to maximum: 0.7) %, and the target mean RBE deviations of the clinical target volumes of 16 patient cases were − 0.4 (− 1.5 to 0.2) %.ConclusionsThe residual RBE difference between RayStation and Syngo was found to be ≤ 1.0%. Thus, we can propose to use RayStation for clinical CIRT treatment planning. However, the potential differences due to the absorbed beam model warrants further exploration.