Objective.Optical computed tomography (CT) is one of the leading modalities for imaging gel dosimeters used in the verification of complex radiotherapy treatments. In previous work, a novel fan-beam optical CT scanner design was proposed that could significantly reduce the volume of the refractive index baths that are commonly found in optical CT systems. Here, the proposed scanner has been manufactured and commissioned.Approach.Image reconstruction is performed through algebraic reconstruction technique and iterated using the fast iterative shrinkage-thresholding algorithm (FISTA) algorithm. Ray tracing for algebraic reconstruction was performed using an in-house developed ray tracing simulator. A set of Sylgard® 184 phantoms were created to commission spatial resolution, geometric deformity, contrast-to-noise ratio (CNR), and scan settings.Main Results.The scanner is capable of a 0.929 mm-1spatial resolution, observed at 200 iterations, although the spatial resolution is highly dependent on the number of iterations. The geometric distortion, measured by scanning a needle phantom with the prototype scanner as well as a conventional x-ray CT was found to be within <0.25 mm. The CNR was found to peak between 65 and 190 occurring between 50 and 100 iterations and was highly dependent on the region chosen for background noise calculation. The proposed scanner is capable of scanning and reading out slices in less than 1 min per slice.Significance.This work displays the viability of a fan-beam optical CT scanner with minimal index matching using ray-traced algebraic reconstruction.