BackgroundThe manual microscopic examination (MME) of urine sediment is labor-intensive, time-consuming, and imprecise. Therefore, automated urinalysis systems based on flow cytometry or digital imaging techniques could replace MME. The purpose of this study was to evaluate the rate of MME using two automated urine sediment analyzers, alone and in combination. MethodsThis study was conducted using the freshly collected urine specimens of 1055 in-patients and 1119 out-patients. All samples were analyzed using UF-1000i (Sysmex Corporation) and Cobas 6500 instrument (Roche Diagnostics International). The rate of MME was evaluated using two analyzers, both individually and in combination. ResultsUsing the UF-1000i alone, 34.2% and 16.8%, respectively, of in- and out-patient samples were analyzed by MME, compared to 15.6% and 3.7%, respectively, using the Cobas 6500. In combined assay using the UF-1000i followed by the Cobas 6500, 27.9% and 11.3% in-patient samples required on-screen review and MME, respectively. And the respective rates were 10.3% and 2.7% of out-patient. Samples using the Cobas 6500 followed by the UF-1000i, 42.3% and 11.3% in-patient needed on-screen review and MME, respectively. And the respective rates were 18.9% and 2.7% of out-patient samples. ConclusionsUse of the Cobas 6500 compared to the UF-1000i resulted in decreases in the rate of MME from 34.2% to 15.6% for in-patient samples, and from 16.8% to 3.7% for out-patient samples. Use of the Cobas 6500 reduced the rate of MME, and compared to use of only the Cobas 6500, the combined use resulted in a reduction in the rate of on-screen review.