We have detected the ultra-low field nuclear magnetic resonance signal from water samples using a high-Tc dc-SQUID sensor. The measurements were carried out in a homemade magnetically shielded room. Resonance spectra of 1H from tap water and other substance samples were obtained in the field range from 7-110μT corresponding to resonance frequency 300-4.68kHz. Two kind of experimental systems were built, the first one is a directly coupled system, its signal to noise ratio in a single-shot measurement is around 4 for about 15 ml water. The second one used a Cu coil to transfer the flux to the SQUID sensor. Signal to noise ratio was improved to about 20 in a single-shot measurement for 5ml water, which benefits from the improvement of coupling efficiency. The effect of residual gradient in the magnetically shielded room was also investigated. J-coupling of 2,2,2-Trifluoroethyl alcohol was measured, the peaks are consistent with high field results.