Development of combined mass spectrometry ionization sources has enabled expansion of the application and scope of mass spectrometry. A novel hybrid ionization system combining vacuum ultraviolet (VUV) and atmospheric pressure chemical ionization (APCI) was constructed. Gaseous samples were self-aspirated into an ionization zone through a capillary by negative pressure, generated by high-speed airflow based on the Venturi effect. Compared with APCI mode alone, the signal-to-noise ratio (S/N) in APCI/VUV mode was increased by about 276-times. To increase the ionization efficiency further, correlated experimental conditions were optimized. Four types of volatile organic compounds (VOCs) were tested to evaluate the performance of the APCI/VUV ion source. Excellent linearity and limit of detection were achieved for compounds in mixed solutions. Quantitative analyses of four VOCs (toluene, cyclohexanone, styrene and ethylbenzene) using APCI/VUV-MS were done, and the relative standard deviations (RSDs) were 1.57%, 6.30%, 4.49% and 8.21%, respectively, indicating that the APCI/VUV ionization source had excellent reproducibility. Our results demonstrated that the developed method was promising for analyzing VOCs as well as being rapid, simple, and easy to operate.