We demonstrate multiplexed readout of 43 transition edge sensor (TES) bolometers operating at 90 mK using a frequency division multiplexing (FDM) readout chain with bias frequencies ranging from 1 to 3.5MHz and a typical frequency spacing of 32kHz. We improve the previously reported performance of our FDM system by two important steps. First, we replace the coplanar wires with microstrip wires, which minimize the cross talk from mutual inductance. From the measured electrical cross talk (ECT) map, the ECT of all pixels is carrier leakage dominated. Only five pixels show an ECT level higher than 1%. Second, we reduce the thermal response speed of the TES detectors by a factor of 20 by increasing the heat capacity of the TES, which allows us to bias all TES detectors below 50% in transition without oscillations. We compare the current-voltage curves and noise spectra of the TESs measured in single-pixel mode and multiplexing mode. We also compare the noise equivalent power (NEP) and the saturation power of the bolometers in both modes, where 38 pixels show less than 10% difference in NEP and 5% difference in saturation power when measured in the two different modes. The measured noise spectrum is in good agreement with the simulated noise based on measured parameters from an impedance measurement, confirming that our TES is dominated by phonon noise.