In recent years, handheld and portable terahertz instruments have been in rapid development for various applications ranging from non-destructive testing to biomedical imaging and sensing. For instance, we have deployed our Portable Handheld Spectral Reflection (PHASR) Scanners for in vivo full-spectroscopic imaging of skin burns in large animal models in operating room settings. In this paper, we debut the polarimetric version of the PHASR Scanner, and describe a generalized calibration technique to map the spatial and spectral dependence of the Jones matrix of an imaging scanner across its field of view. Our design is based on placement of two orthogonal photoconductive antenna (PCA) detectors separated by a polarizing beam splitter in the PHASR Scanner housing. We show that as few as three independent measurements of a well-characterized polarimetric calibration target are sufficient to determine the polarization state of the incident beam at the sample location, as well as to extract the Jones propagation matrix from the sample location to the detectors. We have tested the accuracy of our scanner by validating polarimetric measurements obtained from a birefringent crystal rotated to various angles, as compared to the theoretically predicted response of the sample. This new version of our PHASR scanner can be used for high-speed imaging and investigation of heterogeneity of polarization-sensitive samples in the field.