In this study, we developed a facile and low-cost method to prepare a modified electrode by coating the pencil graphite electrode (PGE) surface with 8-nitroquinoline (8-NQ). The surface modification process was carried out by electrochemical reduction of 8-NQ on the PGE surface in ethanol solution. The optimum conditions were determined for the experimental studies. The differential pulse voltammetry was used to determine rutin with the modified electrode (8-NQ/PGE). The developed electrode showed excellent performance for rutin oxidation in a pH 3.0 Britton Robinson buffer (BR). In this media, linearly increasing anodic peak currents were observed with the concentration of rutin in two wide concentration ranges (i.e., 0.016–0.96 μM and 0.96‒19.73 μM), the sensor showed a low detection limit (i.e., 4.14 nM) (3 s m−1). It displayed good stability and selectivity. Also, it was used successfully for the determination of rutin in orange juice samples.