The oxide layer of a titanium surface is very stable, and seems to result in excellent biocompatibility and successful osseointegration. The purpose of this study was to investigate the effects of high anodic oxidation voltages on the surface characteristics of titanium implants and the biologic response of rabbit tibiae. Bone tissue responses were evaluated by removal torque tests and histomorphometric analysis. Screw-shaped implants with microthreads were made of commercially pure titanium (Grade II). We prepared anodized implants under 300 V (group I), 400 V (group II), 500 V (group III) and 550 V (group IV). The surface characteristics of specimens were inspected according to three categories: surface morphology, surface roughness and oxide layer thickness. The screw-shaped implants were installed in rabbit tibiae. The removal torque values were measured and histomorphometric analysis was done after 1- and 3-month healing periods. Data indicate that as anodic oxidation voltage increased above 300 V, oxide layer thickness increased rapidly and pore size also increased. The roughness values of the implants increased with voltage up to 500 V, but decreased at 550 V. In the removal torque test, group III showed higher values than groups I and II at a statistically significant level (P < 0.05) after a 1-month healing period. In histomorphometric analysis, groups III and IV, after a 3-month healing period, showed greater bone to implant contact ratios for the total implant surface than did group I (P < 0.05).