We introduced an advanced piezoelectric characterization method with mechanical excitation on partial electrode samples, which can determine additional physical parameters the conventional full-electrode electrical excitation method cannot provide. A non-electrode sample with only 10% electrode at the center of a rectangular k31 piezoelectric plate has the benefits for measuring the extensive parameters directly and much precisely. In this paper, we derived first the exact analytical solutions on the partial electrode configuration, including the 10% mechanical excitation part in a partial non-electrode sample for calculating there sonance vibration mode and mechanical quality factors of the composite bar sample for obtaining the extensive elastic compliance and loss factor. The result suggests that the almost accurate values can be obtained even for a specimen with up to 40% center electrode area in the k31 mode samples. Second objective is to compare three different simulations (ATILA FEM, 6-terminal Equivalent Circuit and Analytical Solution). This mechanical excitation method has the benefit in measuring physical parameters which cannot be obtained directly from the IEEE Standard electrical excitation method. Keywords: Direct loss measurement; intensive and extensive parameters; resonance vibration mode; PZT; mechanical quality factor;