ObjectiveWe investigated how a commercially available smartwatch that measures peripheral blood oxygen saturation (SpO2) can detect hypoxemia compared to a medical-grade pulse oximeter.MethodsWe recruited 24 healthy participants. Each participant wore a smartwatch (Apple Watch Series 6) on the left wrist and a pulse oximeter sensor (Masimo Radical-7) on the left middle finger. The participants breathed via a breathing circuit with a three-way non-rebreathing valve in three phases. First, in the 2-minute initial stabilization phase, the participants inhaled the ambient air. Then in the 5-minute desaturation phase, the participants breathed the oxygen-reduced gas mixture (12% O2), which temporarily reduced their blood oxygen saturation. In the final stabilization phase, the participants inhaled the ambient air again until SpO2 returned to normal values. Measurements of SpO2 were taken from the smartwatch and the pulse oximeter simultaneously in 30-s intervals.ResultsThere were 642 individual pairs of SpO2 measurements. The bias in SpO2 between the smartwatch and the oximeter was 0.0% for all the data points. The bias for SpO2 less than 90% was 1.2%. The differences in individual measurements between the smartwatch and oximeter within 6% SpO2 can be expected for SpO2 readings 90%–100% and up to 8% for SpO2 readings less than 90%.ConclusionsApple Watch Series 6 can reliably detect states of reduced blood oxygen saturation with SpO2 below 90% when compared to a medical-grade pulse oximeter. The technology used in this smartwatch is sufficiently advanced for the indicative measurement of SpO2 outside the clinic.Trial RegistrationClinicalTrials.gov NCT04780724