We investigated the feasibility of fast data acquisition in heavy ion CT (IonCT) technique with an X-ray intensifying screen-charged coupled device (CCD) camera system. This technique is based on measuring the residual range distribution of heavy ions after passing through an object. We took a large number of images with a CCD camera for one projection by changing the range shifter (RS) thickness to obtain a characteristic curve similar to a Bragg curve and then to estimate the relative residual range. We used a high quality Electron Multiplying CCD (EMCCD) camera, which drastically reduced data acquisition time. We also used a parallel-plate ionization chamber upstream of an object to monitor the time variation in heavy ion beam intensity from a synchrotron accelerator and to perform beam intensity correction for all EMCCD images. Experiments were conducted using a broad beam of <sup xmlns:mml="http://www.w3.org/1998/Math/MathML" xmlns:xlink="http://www.w3.org/1999/xlink">12</sup> C, which was generated by spreading out the pencil beam accelerated up to 400 MeV/u by the Heavy Ion Medical Accelerator, in Chiba (HIMAC) at the National Institute of Radiological Sciences, with a scatterer. We demonstrated that a fast CT data acquisition, 14 min for 256 projections, is possible for an electron density phantom, consisting of six rods with a relative electron density resolution of 0.017, using the proposed technique with HIMAC.