BackgroundTransit time flow measurement (TTFM) is widely used in coronary artery bypass grafting (CABG); however, its predictive value is unclear. We aimed to identify new factors to evaluate graft quality using fast Fourier transform (FFT).MethodsIntraoperative and postoperative 2-year follow-up data of 114 patients undergoing CABG from January 2017 to December 2018 were collected. The TTFM waveform was transformed by FFT. Mean graft flow (MGF), pulse index, the amplitude of the main wave in FFT (H0), the amplitude of the first harmonic (H1), H0/H1, and the frequency of the first harmonic (P) were analyzed as predictors using logistic regression and receiver operating characteristic (ROC) curves.ResultsThe overall graft patency rate was 80.3%, and the incidence of major adverse cardiac and cerebrovascular events (MACCEs) was 14.9%. The results demonstrate that compared with the graft failure group, MGF, H0, and H1 were higher, but H1 and P were lower in the patent group. With univariate and multivariate logistic regression analyses, the decrease in H0 and H1 and the increase in P were independent risk factors for graft failure, while the decrease in MGF and the increase in H0/H1 were only statistically significant with a univariate analysis. In the cardiovascular events group, the increase in P was an independent risk factor. With a ROC curve analysis, MGF, H0, H1, H0/H1, and P predicted graft failure, while only P predicted cardiovascular events. None of the indicators showed predictive value for MACCEs.ConclusionsTTFM waveforms after FFT can be used to evaluate graft quality and cardiovascular events, but have no predictive value for MACCEs.