Plastic scintillation detectors are a new instrument of stereotactic photon-beam dosimetry. The clinical application of the plastic scintillation detector Exradin W1 at the Siemens Artiste and Elekta Synergy accelerators is a matter of current interest. In order to reduce the measurement uncertainty, precautions have to be taken with regard to the geometrical arrangement of the scintillator, the light-guide fiber and the photodiode in the radiation field. To determine the “Cerenkov light ratio” CLR with a type A uncertainty below 1%, the Cerenkov calibration procedure for small-field measurements based on the two-channel spectral method was used. Output factors were correctly measured with the W1 for field sizes down to 0.5×0.5cm2 with a type A uncertainty of 1.8%. Measurements of small field dose profiles and percentage depth dose curves were carried out with the W1 using automated water phantom profile scans, and a type A uncertainty for dose maxima of 1.4% was achieved. The agreement with a synthetic diamond detector (microDiamond, PTW Freiburg) and a plane parallel ionization chamber (Roos chamber, PTW Freiburg) in relative dose measurements was excellent. In oversight of all results, the suitability of the plastic scintillation detector Exradin W1 for clinical dosimetry under stereotactic conditions, in particular the tried and tested procedures for CLR determination, output factor measurement and automated dose profile scans in water phantoms, have been confirmed.