It has been shown that lead-free and lead-composite aprons do not perform as well as their 100% lead counterparts, and secondary fluorescent radiation can contribute to staff doses. The aim of this project is to measure the transmission through the current cohort of aprons in use in our Cardiology and Interventional Radiology departments, to examine the effectiveness of their protection and determine the optimal transmission and design of protective aprons for future purchasing. Apron designs include full length, backless, skirt/vest aprons, both single panel and overlap. Measurements performed in narrow, broad and the recommended inverse-broad beam geometries (I.S.EN61331-1-2014) are compared with a 100% lead apron for a number of beam qualities. Aprons tested include Xenolite TB and NL, Safety First, Rayshield and Infab lead-free and bi-layer aprons, as well as a 100% lead Amray apron. Calibrated Harshaw TLD-100 and TLD-100H dosimeters were placed on the aprons of several members of staff, including an IR radiologist and a Cardiology registrar, to measure the effectiveness of the apron at multiple sites in a clinical setting. Preliminary results show that for high throughput/high dose procedures, lead equivalence of 0.25 mm is inadequate. The extent of the overlap is not sufficient on some models of apron, particularly on the side of the apron closest to the tube. The suitability of backless aprons will also be discussed. Results will inform minimum requirements, such as the type and thickness of barrier material, apron design and extent of the overlap required for optimal protection.