Introduction. Dispersion in an accelerator quantifies the deviation of the proton trajectory when there is a momentum deviation. We present for the first time a safe method of measuring dispersion in the clinic, using a scintillator detector and the momentum deviations within a spill. This is an important accelerator quantity as we found that this is the reason behind the large dose fluctuation in our absolute dosimetry measurement. Methods. Dispersions are measured for nine energies in a Hitachi ProBeat system at three locations (isocenter and at two profile monitors) and at two gantry angles (0 and 90 degrees) by first measuring the spot position and momentum drift within a spill. The spot position drift is measured by the XRV-4000 at the isocenter, and by the two profile monitors located at 0.57 and 2.27 m from the isocenter. The momentum drift is calculated from the intra-spill range drift which is measured using the Ranger accessory. The dispersion at isocenter and its gradient are calculated using the weighted least square regression on the measured dispersions at the three locations. A constraint is formulated on the dispersion and its gradient to ensure minimal intra-spill spot position deviation around the isocenter. Results. The measured intra-spill range and spot positional drift at isocenter are less than and respectively. The momentum spread calculated from the range drift are less than 0.08%. The dispersion at the isocenter ranged from to and the zero-crossing happens upstream of isocenter for all energies. 2 of the 9 energies (168.0 and 187.5 MeV) violated the constraint and has an intra-spill spot positional deviation greater than within from the isocenter. Conclusion. This measurement is recommended as part of commissioning and annual quality assurance for accelerator monitoring and to ensure intra-spill spot deviations remain low.