The main objective of this study is to evaluate the low-cost, open-source HEGduino device as a tissue oximetry monitor to advance the research of somatic NIRS monitoring. Specifically, this study analyzes the use of this portable functional NIRS system for detecting the cessation of blood flow due to vascular occlusion in an upper limb.19 healthy patients aged between 25 and 50 were recruited and monitored using HEGduino device. Participants underwent a vascular occlusion test on one forearm. Raw values collected by HEGduino as well as the processed variables derived from the measurements were registered. Additional variables to characterize the signal noise during the tests were also recorded.The results of the data distribution curves for all the subjects in the study accurately detected the physiological events associated with transient tissue ischemia. The statistical analysis of the recorded data showed that the difference between the baseline values recorded by the red led (RED) and its normalized minimum variable was always different from zero (p < 0.014). Furthermore, the difference between the normalized baseline values recorded by the infrared led (IR) and the corresponding normalized minimum value was also different from zero (p < 0.001). The R-squared coefficient of determination for the noise variables considered in this study on the normalized RED and IR values was 0.08 and 0.105, respectively.The study confirms the potential of HEGduino system to detect an interruption of the blood flow by means of variations in regional tissue oxygen saturation. This study demonstrates the potential of the HEGduino device as a monitoring alternative to advance the study of the applicability of NIRS in muscle tissue oximetry.
Read full abstract