The energy responses for the KLT-300(LiF:Mg,Cu,Na,Si, Korea), GR-200(LiF:Mg,Cu,P, China) and MCP-N(LiF:Mg,Cu,P, Poland) thermoluminescence(TL) pellets were studied for a photon radiation with energies from 1.25 MeV(60Co) to 21 MV (Microtron) to verify the usefulness of the calibration for the radiotherapy beams. The International Atomic Energy Agency (IAEA) and the World Health Organization (WHO) have performed thermoluminescence dosimetry (TLD) audits to verify the calibration of the beams by TL powder, but TL pellets were used in this study because the element correction factor (ECF), defined as the factor to correct the variations that all TL dosemeters cannot be manufactured to have exactly the same TL efficiency, for each TL pellet could be accurately derived and be handled conveniently when compared with the powder. Also several works for the energy response of the TLDs were done for the low-energy photon beams up to 60Co, but they will be extended in this experiment to the high photon energies (up to 20 MV), which are widely used in the therapy level of a radiation. The PTW 30006 ionisation chamber was calibrated by the Korea primary standards to establish the air-kerma rates and the TL pellets were irradiated in a specially designed waterproof pellet holder in a water phantom (30 x 30 x 30 cm3) just like the IAEA postal audits programme. This result was compared with that of another type of phantom [10 (W) x 10 (L) x 10 (H) cm3 PMMA Perspex phantom for the 60Co and 6 MV photon, and 10 x 10 x 20 (H) cm3 for the 10 and 21 MV photon] for its convenient use and easy handling and installation in a hospital. The results show that the differences of the responses for the water phantom and PMMA Perspex phantom were negligible, which is contrary to the general conception that a big difference would be expected. For an application of these results to verify the therapy beams, an appropriate energy correction factor should be applied to the energies and phantom types in use.
Read full abstract