We present an innovative, reliable, and antibody-free analytical method to determine multiple intact natriuretic peptides in human plasma. These biomolecules are routinely used to confirm the diagnosis and monitor the evolution of heart failure, so that their determination is essential to improve diagnosis and monitor the efficacy of treatment. However, common immunoassay kits suffer from main limitations due to high cross-reactivity with structurally similar species. In our method, we pre-treated the sample by combining salting-out with ammonium sulfate with microextraction by packed sorbent technique. Analyses were then carried out by ultra-high performance liquid chromatography-electrospray ionization-tandem mass spectrometry. The use of 3-nitrobenzyl alcohol as a supercharger reagent enhanced the ESI ionization and improved the signal-to-noise ratio. The analytical protocol showed good linearity over one order of magnitude, recovery in the range of 94–105 %, and good intra- and inter-day reproducibility (RSD<20 %), and the presence of a matrix effect. Limits of detection were in the range of pg/mL for all peptides (0.2–20 pg/mL). Stability study in plasma samples demonstrated that proper protease inhibitors need to be included in blood collection tubes to avoid peptide degradation. Preliminary analyses on plasma samples from heart failure patients allow the quantification of ANP 1–28 as the most abundant species and the detection of ANP 5–28, BNP 1–32, and BNP 5–32. The method could be used to investigate how cross-reactivity issues among structurally similar species impact determinations by ELISA kits.