Using a pediatric head phantom constructed in our department, we examined a method to reduce exposure by using organ-effective modulation (OEM; Toshiba Medical Systems Corporation, Tochigi) to tilt the gantry during pediatric head computed tomography (CT) scanning. The radiation reduction and CT image standard deviation (SD) were measured at gantry angles at which the orbit was slightly irradiated, partially irradiated, and completely irradiated. The OEM incident surface dose reduction rate was measured using an automatic exposure control (AEC) phantom with a diameter of 6-18 cm. The lens surface dose reduction rate using OEM was 21.2%. When the gantry was tilted and the orbit was completely out of the scanning range, the rate of reduction was 47.8%. OEM incident surface dose reduction rates were 27.4% for a phantom diameter of 18 cm, 22.0% for that of 16 cm, 17.8% for that of 14 cm, 17.2% for that of 12 cm, 8.4% for that of 10 cm, and 0% for that of 8 cm and 6 cm. OEM effectiveness decreased with decreasing phantom diameter. The use of OEM increased the rate of change of SD by 1.25´ when the gantry inclination was 0°, 1.27´ when the gantry inclination was 10°, and 1.27´ when the gantry inclination was 20°in the 12 o'clock position. The degree of reduction in exposure dose to the lens in pediatric head CT imaging was 47.8% by completely removing the lens from the irradiation range using gantry tilt and 21.2% by using OEM. The effect of OEM changed in proportion to tube current. The exposure reduction effect of the OEM decreases with decreasing head size, indicating its reduced effectiveness in head CT scans of smaller infants.