The Really ProMRI study evaluates magnetic resonance imaging (MRI) access for patients with cardiac implantable electronic devices (CIEDs) as well as the performance of magnetic resonance (MR)-conditional leads when undergoing MRI. Patients either with an MR-conditional pacemaker or implantable defibrillator (ICD) system or with at least a component (device or one or more leads) from an MR-conditional system, were asked to fill in a questionnaire when they were referred to a MR scan. The rate of prescription, denial, or execution of MR examinations was evaluated in a 1-year follow-up visit. In total, 555 patients [median age (interquartile range) 72.2 (62.2-78.6); 72% male] were enrolled, 49% (270) with a pacemaker, 51% (285) with an ICD system. Five-hundred and ten patients completed the follow-up period. A total of 37 MRI referrals were reported in 35 patients, with a consequent event rate of 7.0/100 patient-years (CI, 4.9-9.7). Fourteen were denied, while 23 [66%; (CI, 48-81%)] were performed. The number of patients with MR referrals was not statistically different between pacemaker and ICD groups (21 vs. 14; P = 0.178). The rate of scans performed was higher in the pacemaker subjects (19/23 vs. 4/14, P = 0.003), while it was similar between patients with or without a complete MR-conditional system (19/30 vs. 4/7, P = 0.606). In this study, we reported a 7.0/100 patient-year's event rate of MR prescriptions in CIED patients. Many examinations were denied, despite MR-conditional systems, especially in ICD patients. Regulatory and cultural changes are needed to allow wider access to MR imaging in CIED patients with MR-conditional systems.