Neuromonitoring devices are often used in traumatic brain injury.The objective of this report is to raise awareness concerning variations in optimal cerebral perfusion pressure (CPPopt) determination using exploratory information provided by two neuromonitoring monitors that are part of research programs (Moberg CNS Monitor and RAUMED NeuroSmart LogO). We connected both monitors simultaneously to a parenchymal intracranial pressure catheter and recorded the pressure reactivity index (PRx) and the derived CPPopt estimates for a patient with a severe traumatic brain injury. These estimates were available at the bedside and were updated at each minute. Using the Bland and Altman method, we found a mean variation of - 3.8 (95% confidence internal from - 8.5 to 0.9) mm Hg between the CPPopt estimates provided by the two monitors (limits of agreement from - 26.6 to 19.1mm Hg). The PRx and CPPopt trends provided by the two monitors were similar over time, but CPPopt trends differed when PRx values were around zero. Also, almost half of the CPPopt estimates differed by more than 10mm Hg. These wide variations recorded in the same patient are worrisome and reiterate the importance of understanding and standardizing the methodology and algorithms behind commercial neuromonitoring devices prior to incorporating them in clinical use.