To evaluate the ability of the DrugSorb™-AntiThrombotic Removal (ATR) haemoadsorption device utilizing porous polymer bead sorbent technology to remove three commonly used antithrombotic drugs from whole blood. We evaluated the removal of apixaban, rivaroxaban, and ticagrelor by the DrugSorb-ATR haemoadsorption device in a benchtop clinical scale model using bovine whole blood. Blood spiked at clinically relevant concentrations of an antithrombotic agent was continuously circulated through a 300-mL DrugSorb-ATR haemoadsorption device at a flow rate of 300mL/min. Drug concentration was monitored over 6h to evaluate drug removal. Results were compared with a control circuit without the haemoadsorption device. Removal rates at 30, 60, 120, and 360 minutes were: apixaban: 81.5%, 96.3%, 99.3% >99.8%; rivaroxaban: 80.7%, 95.1%, 98.9%, >99.5%; ticagrelor: 62.5%; 75%, 86.6%, >95% (all P<0.0001 vs. control). Blood pH and haematological parameters were not significantly affected by the DrugSorb-ATR haemoadsorption device when compared with the control circuit. DrugSorb-ATR efficiently removes apixaban, rivaroxaban, and ticagrelor in a clinical-scale benchtop recirculation circuit with the bulk of removal occurring in the first 60 minutes. The clinical implications of these findings are currently investigated in patients undergoing on-pump cardiothoracic surgery in two US pivotal trials (ClinicalTrials.gov Identifiers: NCT04976530 and NCT05093504).