This study introduces a practical and cost-effective method for tracking diltiazem (DLZ) analytically. It utilizes a fluorimetric approach that relies on the modulation of fluorescence intensity of a dye called erythrosine B. Through a one-pot experiment performed in an acidic environment, a complex is rapidly formed between DLZ and erythrosine B. By observing the decrease in erythrosine B emission, a linear calibration plot is established, enabling the detection and quantification of DLZ concentrations ranging from 40 to 850 ng/ml. The estimated limits of detection and quantitation were 10.5 and 32.1ng/ml, respectively. The variables affecting the DLZ-dye complex system were carefully adjusted. The validity of the approach was confirmed through a thorough evaluation based on the criteria set by ICH guidelines. The accuracy and precision of the methodology were evaluated, and the standard deviation and relative standard deviation were below 2. The strategy was successfully employed to analyze DLZ in tablets and capsules, and no significant variation between the proposed and reported methods as the values of the estimated t-test and F-test at five determinations were below 2.306 and 6.338, respectively. Notably, the method adheres to the principle of green chemistry by utilizing distilled water as the dispersing medium.