The goal of this research was to study the widely used criteria in Iraq and the suggested criteria from the American association of physics in medicine (AAPM). The breast cancer patients treated at Al-Andalus private hospital underwent a CT simulation. The data were exported to Monaco 5.1 treatment planning system for planning purposes using the intensity-modulated radiation therapy technique with X-ray energies of 6 or 10 MV. The planning was transferred to Agility linear accelerator for irradiation to Octavius-4D phantom detector for pre-treatment verification. The results show that there is no significant difference between the prescribed and measured point dose. In addition, the dose difference is acceptable and there was no significant correlation between the dose error and the result of gamma passing rates for all criteria. A high significant difference appeared between the 3%/2 mm and 3%/3 mm criteria for both local and global gamma passing rates at the two studied thresholds 5% and 10%. All the values of 10% thresholds were highly significant than 5% in local %GP but not significant for the global %GP. The number of segments and the total number of monitor units show an inverse relationship with gamma passing rates. The results clearly show that criteria and thresholds depend on the situation and health of the patient. This means the criteria are to be set at 3%/2 mm for plans which use high doses to organs at risk, while 3%/3 mm may be used for simpler plans with lower doses to organs at risk.