ObjectiveTo statistically analyze the results obtained from biomechanical tests on fixation of femoral neck fractures of Pauwels III type, in synthetic bone, using the dynamic hip system with an anti‐rotation screw, versus a control group. MethodsTen synthetic bones from a Brazilian manufacturer (model C1010) were used, divided into two groups: test and control. In the test group, fixation of an osteotomy was performed with 70° of inclination at the level of the femoral neck, using DHS with an anti‐rotation screw. The resistance of this fixation was evaluated, along with its rotational deviation at 5mm of displacement (phase 1) and at 10mm of displacement (phase 2), which was considered to be failure of synthesis. In the control group, the models were tested in their entirety until femoral neck fracturing occurred. ResultsThe test values in the test group (samples 1 to 5) in phase 1 were: 1,512N, 1,439N, 1,205N, 1,251N and 1,273N, respectively (mean=1,336N; standard deviation [SD]=132N). The rotational deviations were: 4.90°, 3.27°, 2.62°, 0.66° and 0.66°, respectively (mean=2.42°; SD=1.81°). In phase 2, we obtained: 2,064N, 1,895N, 1,682N, 1,713N and 1,354N, respectively (mean=1,742N; SD=265N). The failure loading values in the control group were: 1,544N, 1,110N, 1,359N, 1,194N and 1,437N, respectively (mean=1,329N; SD=177N). The statistical analysis using the Mann‐Whitney test showed that the test group presented maximum loading at a displacement of 10mm, i.e. significantly greater than the failure loading of the control group (p=0.047). ConclusionThe mechanical resistance of the test group was significantly greater than that of the control group.