Mass analysis in an ion trap is conventionally realized through time domain analysis of the ejected ion current collected from an electron multiplier (EM), in which the ion ejection time is found to have a correlation with the mass-to-charge (m/z) ratio of the ion. In this study, we investigated a new method for mass analysis by examining ion ejection signals in the frequency domain. Theoretical analysis and ion trajectory simulations show that ions of the same m/z ratio are ejected from an ion trap at regular intervals, producing a periodic pulsed signal on the EM. The period of this pulsed ejection signal is directly linked to the m/z values of the ions. To realize this method experimentally, a broadband preamplifier was built and integrated on a miniature ion trap mass spectrometer (the "Brick" series from Nier Inc.) to capture this pulsed ion ejection signal collected from the EM. Experimental results were in good agreement with theoretical and simulation analyses. This method has the potential to improve the mass resolution of an ion trap mass analyzer. As a proof-of-concept demonstration, a peak width of 0.1 Da at a m/z value of 281 was achieved in experiments.