Previous reports describing sample stacking on microchip capillary electrophoresis (microCE) have regarded the microchip channels as a closed system and treated the bulk flow as in traditional capillary electrophoresis. This work demonstrates that the flows arising from the intersection should be investigated as an open system. It is shown that the pressure-driven flows into or from the branch channels due to bulk velocity mismatch in the main channel should not be neglected but can be used for liquid transportation in the channels. On the basis of these concepts, a sample preconcentration scheme was developed in a commercially available single-cross glass chip for microCE. Similar to field-amplified stacking injection in traditional CE, a low conductivity sample buffer plug was introduced into the separation channel immediately before the negatively charged analyte molecules were injected. The detection sensitivity was improved by 94-, 108-, and 160-fold for fluorescein-5-isothiocyanate, fluorescein disodium, and 5-carboxyfluorescein, respectively, relative to a traditional pinched injection. The calibration curves for fluorescein and 5-carboxyfluorescein demonstrated good linearity in the concentration range (1-60 nM) investigated with acceptable reproducibility of migration time and peak height and area ratios (4-5% RSD). This preconcentration scheme will be of particular significance to the practical use of microCE in the emerging miniaturized analytical instrumentation.