TLDs dosimeters are frequently presented as a viable choice for dosimetric studies when dealing with mixed neutron-gamma radiation fields. However, this choice is not without some drawbacks, because not only TLD response is highly dependent on particle type but also on neutron energy spectrum. Therefore, a correct screening and calibration of the dosimeter are required, and a simple shift from gamma screening methodology for mixed field is not suitable. This paper presents the design, fabrication and tests of an irradiator for TLD screening for neutron dose measurement using an AmBe source and polyethylene as moderator material. The design of the irradiator was conducted through Monte Carlo simulations using the MCNP5 code. The experimental validation and tests were performed using Indium activation foils and TLD 600 dosimeters. The manufactured irradiator demonstrated to be suitable for TLD screening under neutron source radiation field, offering very good homogeneity conditions in the radiation field so to guarantee same radiation dose delivered to the TLDs.
Read full abstract