The application of portable chromatography–mass spectrometer (GC–MS) is restrained by its detection limits without the development of proper sample pre-concentration methods. The primary focus of this paper is to introduce a practical field measurement methodology for the analysis of volatile organic compounds (VOCs) in soil vapour and groundwater using a portable gas (GC–MS)system for application to in situ assessment of vapour intrusion from VOC contamination. A solid-phase micro-extraction (SPME) technique was applied for sample pre-concentration before the GC–MS measurement. Practical in-field soil gas SPME sampling methods have been developed to optimise the SPME extraction efficiency to then ultimately improve the detection limits of portable GC–MS. An Australian site impacted by a chlorinated VOC, trichloroethylene (TCE), was the subject of the case study. To rapidly assess soil vapour samples in subsurface soil, in-house-developed retractable soil vapour sampling probes (SVSPs) were installed at the site in clusters at depths of 1 m, 2 m and 3 m below ground level at each sampling location. Use of the SVSPs for sampling enabled the generation of a three-dimensional map and distribution contours for TCE concentrations using the in situ measurement results of a portable GC–MS analysis for vapour intrusion investigation. The results of the portable GC–MS analysis were compared with the results from conventional USEPA methods, such as TO-15 and Method 8265 for soil vapour and groundwater samples, respectively. This work demonstrates that the developed methodology of using a portable GC–MS system has the capability for in-field quantitative analysis of VOCs for rapid contaminated site vapour intrusion assessment.