ObjectiveTemperature controlled T1 and T2 relaxation times are measured on NiCl2 and MnCl2 solutions from the ISMRM/NIST system phantom at low magnetic field strengths of 6.5 mT, 64 mT and 550 mT.Materials and methodsThe T1 and T2 were measured of five samples with increasing concentrations of NiCl2 and five samples with increasing concentrations of MnCl2. All samples were scanned at 6.5 mT, 64 mT and 550 mT, at sample temperatures ranging from 10 °C to 37 °C.ResultsThe NiCl2 solutions showed little change in T1 and T2 with magnetic field strength, and both relaxation times decreased with increasing temperature. The MnCl2 solutions showed an increase in T1 and a decrease in T2 with increasing magnetic field strength, and both T1 and T2 increased with increasing temperature.DiscussionThe low field relaxation rates of the NiCl2 and MnCl2 arrays in the ISMRM/NIST system phantom are investigated and compared to results from clinical field strengths of 1.5 T and 3.0 T. The measurements can be used as a benchmark for MRI system functionality and stability, especially when MRI systems are taken out of the radiology suite or laboratory and into less traditional environments.