This study sought to examine the effect of the return electrode's surface area on bipolar RFA lesion size. Bipolar radiofrequency ablation (RFA) is typically performed between 2 3.5-mm tip catheters serving as active and return electrodes. We hypothesized that increasing the surface area of the return electrode would increase lesion dimensions by reducing the circuit impedance, thus increasing the current into a larger tissue volume enclosed between the electrodes. In step 1, exvivo bipolar RFA was performed between 3.5-mm and custom-made return electrodes with increasing surface areas (20, 80, 180mm2). In step 2, exvivo bipolar RFA was performed between 3.5-mm and 3.5-mm or 8-mm electrode catheters positioned perpendicular or parallel to the tissue. In step 3, invivo bipolar RFA was performed between 3.5-mm and either 3.5-mm or 8-mm parallel electrode at the: 1) left ventricular summit; 2) interventricular septum; and 3) healed anterior infarction. In step 1, increasing the surface area of the return electrode resulted in lower circuit impedance (R =-0.65; P< 0.001), higher current (R =+0.80; P< 0.001), and larger lesion volume (R =+0.88; P< 0.001). In step 2, an 8-mm return electrode parallel to tissue produced larger and deeper lesions compared with a 3.5-mm return electrode (P=0.014 and P = 0.02). Similarly, in step 3, compared with a 3.5-mm, bipolar RFA with an 8-mm return electrode produced larger (volume: 1,525 ± 871mm3 vs 306 ± 310mm3, respectively; P< 0.001) and more transmural lesions (88% vs 0%; P< 0.001). Bipolar RFA using an 8-mm return electrode positioned parallel to the tissue produces larger lesions in comparison with a 3.5-mm return electrode.