The aim of this study was to compare two types of light irradiation devices for antimicrobial photodynamic therapy (aPDT). A 660-nm light-emitting diode (LED) and a 665-nm laser diode (LD) were used for light irradiation, and 0.1 mg/L TONS 504, a cationic chlorin derivative, was used as the photosensitizer. We evaluated the light attenuation along the vertical and horizontal directions, temperature rise following light irradiation, and aPDT efficacy against Staphylococcus aureus under different conditions: TONS 504 only, light irradiation only, and TONS 504 with either LED (30 J/cm2) or LD light irradiation (continuous: 30 J/cm2; pulsed: 20 J/cm2 at 2/3 duty cycle, 10 J/cm2 at 1/3 duty cycle). Both LED and LD light intensities were inversely proportional to the square of the vertical distance from the irradiated area. Along the horizontal distance from the nadir of the light source, the LED light intensity attenuated according to the cosine quadrature law, while the LD light intensity did not attenuate within the measurable range. Following light irradiation, the temperature rise increased as the TONS 504 concentration increased in the order of pulsed LD < continuous LD < LED irradiation. aPDT with light irradiation only or TONS 504 only had no antimicrobial effect, while aPDT with TONS 504 under continuous or pulsed LD light irradiation provided approximately 3 log reduction at 30 J/cm2 and 20 J/cm2 and approximately 2 log reduction at 10 J/cm2. TONS 504-aPDT under pulsed LD light irradiation provided anti-microbial effect without significant temperature rise.