Dose verification for a gimbal-mounted image-guided radiotherapy system, Vero4DRT (Mitsubishi Heavy Industries Ltd., Tokyo, Japan) is usually carried out by pretreatment measurement. Independent verification calculations using Monte Carlo methods for Vero4DRT have been published. As the Clarkson method is faster and easier to use than measurement and Monte Carlo methods, we evaluated the accuracy of an independent calculation verification program and its feasibility as a secondary check for Vero4DRT. Computed tomography (CT)-based dose calculation was performed using a modified Clarkson-based algorithm. In this study, 120 patients' treatment plans were collected in our institute. The treatments were performed using conventional irradiation for lung and prostate, 3-dimensional (3D) conformal stereotactic body radiotherapy (SBRT) for the lung, and intensity-modulated radiation therapy (IMRT) for the prostate. Differences between the treatment planning system (TPS) and the Clarkson-based independent dose verification software were computed, and confidence limits (CLs, mean ± 2 standard deviation %) for Vero4DRT were compared with the CLs for the C-arms linear accelerators in the previous study. The results of the CLs, the conventional irradiation, SBRT, and IMRT showed 2.2 ± 3.5% (CL of the C-arms linear accelerators: 2.4 ± 5.3%), 1.1 ± 1.7% (−0.3 ± 2.0%), 4.8 ± 3.7% (5.4 ± 5.3%), and −0.5 ± 2.5% (−0.1 ± 3.6%) differences, respectively. The dose disagreement between the TPS and CT-based independent dose verification software was less than the 5% action level of American Association of Physicists in Medicine (AAPM) Task Group 114 (TG114). The CLs for the gimbal-mounted Vero4DRT were similar to the deviations for C-arms linear accelerators.