Starting in 2029, the High Luminosity LHC (HL-LHC) at CERN is projected to achieve unparalleled instantaneous luminosity (7.5 × 1034 cm-2s-1) for an integrated luminosities of 4500 fb-1. To cope with the extreme conditions of 200 collisions per bunch crossing (pile-up), both on-detector and off-detector electronics of the CMS Electromagnetic Calorimeter (ECAL) will be upgraded. The new system will incorporate a dual-gain trans-impedance amplifier alongside an Application-Specific Integrated Circuit (ASIC) that includes dual 160 MHz ADC channels and features for gain selection and data compression. The lead tungstate crystals and avalanche photodiodes (APDs) will remain in service due to their proven reliability. To counteract the noise amplification in the APDs, attributed to radiation-induced dark currents, the operational temperature of the ECAL will be reduced from 18°C to roughly 9°C. Preliminary tests of these new electronics have yielded promising results. During tests beams at the CERN H4 beamline using an electron beam, the new electronics met both required energy resolution and linearity and achieved a timing resolution of 30 ps for energies above 50 GeV.