Proton therapy requires range verification in order to exploit its full potential. One of the most promising approaches is to monitor prompt gamma-rays produced by nuclear interactions of the therapeutic particles in the patient tissues. A detector with a wide energy range from 100 keV to 15 MeV and excellent time resolution is required to achieve millimetric precision in proton range. During patient treatment, the detector count rates are usually above 106 s-1 and the fraction of pile-up events is very high for commonly used fast inorganic scintillators. We are investigating a full acceptance approach with increased granularity in order to reduce the size of the scintillators and consequently the count rate per channel. Stacking the scintillators in matrices requires suitable multi-channel photo-multipliers and a fitting acquisition system. Here, we present two geometries of CeBr3 crystals 5×5×20 mm3 and 10×10×30 mm3, together with modern silicon photo-multipliers (SiPM) adapted to work with the PETsys TOFPET2 ASIC. The TOFPET2 ASIC was developed for Time-of-Flight Positron Emission Tomography (TOF-PET) applications. Here we show its potential for higher gamma-ray energies and future hybrid imaging. First results of energy resolution of 6.1%–7.8% are achieved at 3.42 MeV using a 241Am9Be source. The time resolution was found to be below 100 ps and studies of the count rates and the dead time of the full system were performed. Different SiPM models are analysed for their impact on the coincidence time resolution.