Global Product Manager Invasive Blood Pressure Monitoring and Interventional Imaging, Smiths Medical International Ltd., Military Road, Hythe, United Kingdom. helen.reeve@smiths-medical.comWe would like to make the following comments in response to the article by Dr. Görlinger et al. (Görlinger K, Kehren CJ, Peters J: Mini-epidemic of erroneous central venous pressure measurements due to malproduction of a specific part of a pressure transducer system. Anesthesiology 2009; 110:1417–8).In 2008, Smiths Medical became aware of a manufacturing fault with its pressure monitoring kits; this fault has since been rectified. The products involved were LogiCal®, NovaTrans®, and TranStar® invasive pressure monitoring kits with trigger flush device, product codes DPSxxxxxx/MX95xxxx/MX96xxxx/MXxxx and SXxxxxxx and manufactured between October 2007 and February 2008. In the affected kits, flow rates were higher than the product specification of 3 ml/h due to communication between the slow-flow orifice and the flush. The cause of the fault was identified as an assembly problem, and corrective actions were implemented to prevent reoccurrence.Customers who had purchased sets manufactured within the specified timeframe were notified of the problem via a field safety notice, and regulatory bodies were informed.Customers were advised to perform a precheck during priming of each device and before connecting to the patient to identify whether the issue was present. Separate instructions to assess kits already in use were also provided. Customers with affected kits were advised to contact their sales representative for free collection and replacement.The following procedure is recommended for testing the flow rates of these kits:Pretest:Devices already in use with patients can be checked by observing the flow rate in the drip chamber as described above. When the flush is not being operated, the flow rate observed in the drip chamber should be:If the flow rate in the drip chamber is greater than this, it should be discarded and replaced with a new device that has passed the pretest.Smiths Medical takes the quality of its products and the satisfaction of its customers very seriously. If a product fails to meet expectations, they should not hesitate to contact us. We investigate all complaints thoroughly and undertake any necessary corrective actions.Global Product Manager Invasive Blood Pressure Monitoring and Interventional Imaging, Smiths Medical International Ltd., Military Road, Hythe, United Kingdom. helen.reeve@smiths-medical.com