PurposeTo reevaluate a tin filter's (TF) dose reduction effect in computed tomography (CT) using a combination of an anthropomorphic chest phantom and a rod-shaped phantom. Methods and materialsA third-generation dual-source CT system equipped with a built-in TF was employed. A chest phantom was scanned under low-dose conditions of 0.2 to 1.0 mGy with the TF at 100 kV (TF100kV) and without it at 100 kV and 120 kV (NF100kV and NF120kV). To eliminate effects other than that of the TF, only filtered back projection (FBP) was used for image reconstruction. On the images of the rod phantom placed inside the lung field, the CT number and the spatial resolution using the modulation transfer function (MTF) were measured. Using these indices plus the noise power spectrum (NPS) that was also measured, the detectability index based on the non-prewhitening model observer (d'NPW) was calculated. ResultsThe CT numbers and MTFs were almost identical across the three conditions. The area under the NPS curve was decreased by 13–17% with the TF compared with non-TF conditions. NPS increases at low frequencies of < 0.06 mm−1 observed in NF120kV and NF100kV were eliminated by TF100kV. The potential dose reduction by the TF, estimated using the d'NPW values, turned out to be 22 to 25%. ConclusionBased on the analysis of the FBP images of a chest phantom, the dose reduction attributable only to the TF was estimated at 22–25%, notably lower than those reported in previous studies.